For used oils, I seek expert guidance to understand testing procedure and interpretaion guideline for the following tests conducted using FTIR

(1) Oxidation
(2) Nitration
(3) Soot
(4) Sulfation

Can any one help me?
 
Posts: 47 | Registered: Mon July 26 2004Reply With QuoteEdit or Delete MessageReport This Post
Bronze Member - 1 or more posts
Posted Hide Post
Kumar,
I think that only nitration extent and soot content can always be read with a good reliability from FTIR spectrum of used oil. Oxidation band near 1700 cm-1 can be superimposed with the band of ester base oil and cannot be sufficiently integrated, and sulfation signals are almost negligible as a consequence of good quality of recent low-sulfur fuels. That is my experience with european motor oils and quality of european fuels.
 
Posts: 6 | Registered: Wed October 20 2004Reply With QuoteEdit or Delete MessageReport This Post
Platinum Member - 50 or more posts
Posted Hide Post
Kumar,

If you e-mail me I will send you an extensive guide for using FTIR.

Greg
 
Posts: 62 | Location: US | Registered: Thu July 01 2004Reply With QuoteEdit or Delete MessageReport This Post
Silver Member - 10 or more posts
Posted Hide Post
Dear Greg,
Pl. send details to me, as I am also interested.
My Email i.d is

dipendra@cetexpetro.com
 
Posts: 24 | Registered: Fri April 23 2004Reply With QuoteEdit or Delete MessageReport This Post
sm
Bronze Member - 1 or more posts
Posted Hide Post
Dear Greg,

I happen to see this discussion going on regarding use of FTIR for used oil analysis.

Please send me details too as I am also interested in participating and enriching my knowledge on the subject.

Incidentally, we have fairly well equipped laboratory for handling used oil analysis as well as other analytical testings, for research projects, pharmaceuticals, etc. complete with AAS, ICP, GC-MS, GCs, HPLCs, FT-IR, UV-VIS, etc. to name a few, at Mumbai in western part of INDIA.

My Email address is : sm@spagroup.net

Regards
 
Posts: 7 | Location: Mumbai, INDIA | Registered: Wed July 21 2004Reply With QuoteEdit or Delete MessageReport This Post
Bronze Member - 1 or more posts
Posted Hide Post
quote:
Originally posted by Greg Livingstone:
Kumar,

If you e-mail me I will send you an extensive guide for using FTIR.

Greg


Dear Greg,

Pl. send details to me, as I am also interested.
My Email i.d is caleen98@yahoo.com

Regards,

Callistus Emeana
 
Posts: 9 | Location: Nigeria | Registered: Mon October 25 2004Reply With QuoteEdit or Delete MessageReport This Post
Gold Member - 25 or more posts
Posted Hide Post
quote:
Originally posted by Kumar:
For used oils, I seek expert guidance to understand testing procedure and interpretaion guideline for the following tests conducted using FTIR

(1) Oxidation
(2) Nitration
(3) Soot
(4) Sulfation

Can any one help me?


I came to know through web search that McGill university, Montreal, QC, Canada has done lot work using FTIR and their findings are being followed by an oil industry in used oil testing I.E. Thermal- lube Quebec ; its web site http://www.thermal-lube.com/
 
Posts: 47 | Registered: Mon July 26 2004Reply With QuoteEdit or Delete MessageReport This Post
Mes
Bronze Member - 1 or more posts
Posted Hide Post
Good morning everybody,
I am new to this forum, in fact i registered because of this thread.

I work in a lubs lab, we did not have any ftir instrument until now, and we have very recently received a perkin-elmer "oil express" analyzer (Spectrum One + autosampler + oilexpress software).

From the first, few, tests we performed it seems that the results (the indexes that are given as results) are not repeatable. I am also not familiar with this "index" format of the results. Perkin Elmer has manuals etc but, the results do not seem to make any sense as I obviously don't know how to interpret them. I really don't know which way I should continue the "research".

So I saw this post
quote:
Kumar,

If you e-mail me I will send you an extensive guide for using FTIR.

Greg

and I thought that, maybe someone could help here! Is there a way that I could get this guide too?

How can I use this new, state of the art, instrument for used oil analysis? Maybe someone has already experience of using the same software too?

Thanks!!!
 
Posts: 3 | Registered: Tue January 04 2005Reply With QuoteEdit or Delete MessageReport This Post
Bronze Member - 1 or more posts
Posted Hide Post
Dear Greg,

I am interesting on this method. Would you please e-mail me guide for FTIR? My email address is Sherry.xia@honeywell.com. Thanks
 
Posts: 6 | Registered: Mon May 15 2006Reply With QuoteEdit or Delete MessageReport This Post
Platinum Member - 50 or more posts
Posted Hide Post
Dear Greg,
If you happen to visit this thread ,please send me the Guide for FTIR at yourminimac@hotmail.com I will be obliged
Prabhakar Agrawal
 
Posts: 93 | Location: Ranchi, India | Registered: Tue August 03 2004Reply With QuoteEdit or Delete MessageReport This Post
Platinum Member - 50 or more posts
Posted Hide Post
quote:
Originally posted by Greg Livingstone:
Kumar,

If you e-mail me I will send you an extensive guide for using FTIR.

Greg


Dear Mr.Greg

My E-mail address is nnkrishnakumar_menon@rediffmail.com.Please help me also by sending that user guide.

Thanks in advance.
 
Posts: 90 | Registered: Thu December 16 2004Reply With QuoteEdit or Delete MessageReport This Post
Silver Member - 10 or more posts
Posted Hide Post
Dear Mes,

We also had a PE FTIR Spectrum One + Sampler + Oil Express Software and we conduct and huge comparison with others parameters like TBN, TAN, filtration in millipore system and WM, and the software present problems of low values of Oxidation and with high concentration of WM, insoluble contaminants, low TBN and high TAN, in our experience the software in some cases don't give good information....
I know there is an ASTM draft about FTIR on condition monitoring and probably during this year will be official...
now patience my friend....

regards
pcancino
 
Posts: 19 | Registered: Tue August 29 2006Reply With QuoteEdit or Delete MessageReport This Post
Mes
Bronze Member - 1 or more posts
Posted Hide Post
Dear pcancino,
At first thank you for replying.

I checked in www.astm.org and it seems that they work on a standard for Water Contamination (WK6542), I guess this is the one you mean. It is mentioned that there will not be any reference to limits in the final standard, but one has to do stats analysis etc. This is logical of course, but what does one do when one receives very few samples from a specific engine? This is an often phenomenon in our lab...

I would be most grateful if you could mention any values for the other parameters that above or below them you consider the results of the software being disputable. If there is no problem with that, you would really help me to limit my own recearch in areas that there is a meaning. For the moment, I feel like our lab has wasted a big amount of time and money by gathering info that does not seem to be useful at all.

Thanks again for your time.
 
Posts: 3 | Registered: Tue January 04 2005Reply With QuoteEdit or Delete MessageReport This Post
Silver Member - 10 or more posts
Posted Hide Post
Mes, give me your email and I’ll send you a copy of the ASTM Draft.

I can't send you data of results due to we didn't implement the DLL from PE to perform the calculation for the same reason that you got problems now, we implemented the DLL from ExxonMobil because we're a third party laboratory.

Regards
pcancino
 
Posts: 19 | Registered: Tue August 29 2006Reply With QuoteEdit or Delete MessageReport This Post
Mes
Bronze Member - 1 or more posts
Posted Hide Post
I would appreciate if you send me the draft.
My email is mkam@mailbox.gr
Thanks in advance
 
Posts: 3 | Registered: Tue January 04 2005Reply With QuoteEdit or Delete MessageReport This Post
Bronze Member - 1 or more posts
Posted Hide Post
I have interested to real know the interpretation of the data from used oil analysis by FTIR, and trend monitoring of the wearcheck
 
Posts: 1 | Registered: Wed January 17 2007Reply With QuoteEdit or Delete MessageReport This Post
Bronze Member - 1 or more posts
Posted Hide Post
My email addres: mh_asim79@yahoo.com please send extensive guide for using FTIR.and also tell me how to measure and calculat the peak of oxidation and nitration
 
Posts: 2 | Location: Karachi | Registered: Fri January 11 2008Reply With QuoteEdit or Delete MessageReport This Post
Silver Member - 10 or more posts
Posted Hide Post
pls send me the mail regarding oil analysis by FTIR.My email add is - neet862007@gmail.com


Thanks a lot.
 
Posts: 10 | Location: india | Registered: Thu February 28 2008Reply With QuoteEdit or Delete MessageReport This Post
 Previous Topic | Next Topic powered by eve community  
 

Noria Corporation    forums.noria.com    Message Boards  Hop To Forum Categories  Oil Analysis    FTIR - Used oil testing


© 2006 Noria Corporation. All Rights Reserved.
Guidelines and Terms

Go to our old message boards.