We have been receiving oil analysis results from the lab showing a flash point of 55 Deg for a gas comp seal oil, however when the platform itself does the test the receive a much higher flash point (88). The lab also states that opening the sample is like opening a bottle of fizzy pop.
If the samples are being tested and collected in the same way how can this be possible? |
||
|
|
Double Platinum Member - 100 or more posts |
Check if you are using the same methods.
Closed Cup vs. Open Cup. |
|||
|
|
Platinum Member - 50 or more posts |
The open cup method is not recommended for those materials whose flash point is less than 75°c as per the ASTM methods.(ASTM D 92).Also this much difference is not possible.
|
|||
|