Dear Folks, we're using an APC from Hyac Royco, and we can't perform a good trend analysis in the ISO VG 680 used in the motorized wheel of Komatsu, we're thinking in implement ISO 4407 but our main concern is how much time need an analist to perform the test? if someone can provide much info, i'll be gratefull.

rgds
pcancino
 
Posts: 19 | Registered: Tue August 29 2006Reply With QuoteEdit or Delete MessageReport This Post
Double Platinum Member - 100 or more posts
Posted Hide Post
Hello.

The preparation time for microscopy for an ISO VG 680 grade oil will take time (I would estimate 1 hour for 100 ml oil).

The counting (ISO 4407) is also time consuming -I would estimate approx. at least 15-30 minutes for an experienced analyst.

I would recommend to look into the LaserNet Fines from Spectro. You would need to dilute the oil with some clean solvent due to the high viscosity.
 
Posts: 204 | Location: Invicta Oil Lab, Norway | Registered: Fri March 05 2004Reply With QuoteEdit or Delete MessageReport This Post
Silver Member - 10 or more posts
Posted Hide Post
Dear Mr. Hughes

In your opinion is better complement ISO 4406 (APC) + Ferrography or ISO 4407 for ISO VG 320 to 680 + ISO 4406 (APC) ISO VG 220 and lowers grades + Ferrography.

Thank you very much.
 
Posts: 19 | Registered: Tue August 29 2006Reply With QuoteEdit or Delete MessageReport This Post
Double Platinum Member - 100 or more posts
Posted Hide Post
quote:
Originally posted by pcancino:
Dear Mr. Hughes

In your opinion is better complement ISO 4406 (APC) + Ferrography or ISO 4407 for ISO VG 320 to 680 + ISO 4406 (APC) ISO VG 220 and lowers grades + Ferrography.

Thank you very much.


I would look into diluting the sample with clean solvent (heptane-n or similar) for thicker oils (due to visc and evt. high contamination levels) and continue with APC+Ferrography in your case.
 
Posts: 204 | Location: Invicta Oil Lab, Norway | Registered: Fri March 05 2004Reply With QuoteEdit or Delete MessageReport This Post
Bronze Member - 1 or more posts
Posted Hide Post
pcancino. If you do decide to dilute your APC's make sure that you run 10-20 samples of a test oil to check your repeatability. You may be dissapointed.
 
Posts: 6 | Registered: Wed March 31 2004Reply With QuoteEdit or Delete MessageReport This Post
Silver Member - 10 or more posts
Posted Hide Post
Make sure you monitor the mag plug capture levels. The studies we did were all using pore blockage PC .
 
Posts: 18 | Location: USA | Registered: Wed May 19 2004Reply With QuoteEdit or Delete MessageReport This Post