Is there any device you know ,that test % oil in water?
>2%oil ( emulsion property)

I'm using acid cracking procedure to split the oil.I want to use new technology that save my time in my lab and get the result as soon as possible, becuase this result is very important in our process.

Appreciate any reply
chemist
 
Posts: 12 | Location: garmco Baharain | Registered: Tue June 22 2004Reply With QuoteEdit or Delete MessageReport This Post
Double Platinum Member - 100 or more posts
Posted Hide Post
The standard screening test for water in lube oil is the "Crackle test" which only requires a hot plate to conduct. Most commercial labs use this test to determine if a further test to quantify the amount of water is required. It is possible to use this test as trigger for water removal actions in your equipment such as filtration with water absorbing elements, centrifuging, or other water removal methods including oil changes.
There are water saturation meters available from several lubrication product vendors. A web search should be able to locate vendors with equipment to fit your needs.
 
Posts: 162 | Location: Midwest, USA | Registered: Tue April 13 2004Reply With QuoteEdit or Delete MessageReport This Post
Double Platinum Member - 100 or more posts
Posted Hide Post
BuckHntr:

% oil in water is VERY different from % water in oil.....

chemist:
I know there are several different instruments. These instruments are often used onsite, controlling waste water. If you give me a day, I will reply with some manufactorers.

In our lab we use FT-IR. The onsite instruments are usually just indicators, but I might find some more advanced systems.
 
Posts: 207 | Location: Invicta Oil Lab, Norway | Registered: Fri March 05 2004Reply With QuoteEdit or Delete MessageReport This Post
Double Platinum Member - 100 or more posts
Posted Hide Post
Here are some suppliers. (Mostly monitors):

Turner Designs: http://www.oilinwatermonitors.com/

Ionics Instruments

Teledyne
 
Posts: 207 | Location: Invicta Oil Lab, Norway | Registered: Fri March 05 2004Reply With QuoteEdit or Delete MessageReport This Post
Double Platinum Member - 100 or more posts
Posted Hide Post
Mr. Hughes
You are absoluly right and I misread your question. Hope you find what your looking for.
 
Posts: 162 | Location: Midwest, USA | Registered: Tue April 13 2004Reply With QuoteEdit or Delete MessageReport This Post
Silver Member - 10 or more posts
Posted Hide Post
Thank you very much I apprecaite each reply

regards
chemist
 
Posts: 12 | Location: garmco Baharain | Registered: Tue June 22 2004Reply With QuoteEdit or Delete MessageReport This Post
Platinum Member - 50 or more posts
Posted Hide Post
Chemist

Is this coolant Concentration you're talking about ?

Are you looking for on site or Off Site equipments?
 
Posts: 72 | Location: INDIA | Registered: Wed May 12 2004Reply With QuoteEdit or Delete MessageReport This Post
Silver Member - 10 or more posts
Posted Hide Post
FRANKY,
yes this is coolant concentration analysis
and I need lab analysis equipment.
I take a two litres sample out of the coolant system every day for analysis.
Its concentration limit is 2.3% to 4.3%
acid cracking procedure takes time and consume acid.
I'm looking for one equipment that save my time, and gives the test results in minutes.

And if you have further information about in site equipment please send it also , I may need in the near future.

Thank you very much
chemist
 
Posts: 12 | Location: garmco Baharain | Registered: Tue June 22 2004Reply With QuoteEdit or Delete MessageReport This Post
Bronze Member - 1 or more posts
Posted Hide Post
Hi Chemist.

Not a device as such, but a method. try solvent extraction of the oil from the water followed by evaporation and gravimetric determination.

Cheap: Sep' funnel, small ammount of non-polar solvent, a balance and a beaker.

2% of 2 litres = 40 grams

You may be able to use much less sample?
 
Posts: 1 | Location: melbourne | Registered: Wed March 08 2006Reply With QuoteEdit or Delete MessageReport This Post