Bronze Member - 1 or more posts |
My SRM for ASTM D4739 is NIST 1848 Lubricating oil additive. It's certified value for TBN is 44~55.2. But my results show negative values.
Other samples results are minus as well. I use Mettler Toledo Autotitrator. Why do I get negative values on TBN? Is there any other SRM available for ASTM D4739 not D2896? Please recommend one. Thanks, Viv This message has been edited. Last edited by: Viv, |
||
|
|
Bronze Member - 1 or more posts |
What is the unit of the TBN?
|
|||
|
|
Silver Member - 10 or more posts |
TBN is measured in mg KOH/g. However, I'd like to know if it is the only SRM for TBN. Is there any others? I'm also looking to check my titrator performance.
|
|||
|
|
Bronze Member - 1 or more posts |
You don't say if this problems is just for this range of testing.
We had a simmilar situation when we started to use a new titrator (replacement) as you know the equipment reads millivolts. What happened, whe inverted the inputs in the rear part of titrator. May be this help. |
|||
|