We have been receiving oil analysis results from the lab showing a flash point of 55 Deg for a gas comp seal oil, however when the platform itself does the test the receive a much higher flash point (88). The lab also states that opening the sample is like opening a bottle of fizzy pop.
If the samples are being tested and collected in the same way how can this be possible?
 
Posts: 1 | Registered: Thu August 04 2005Reply With QuoteEdit or Delete MessageReport This Post
Double Platinum Member - 100 or more posts
Posted Hide Post
Check if you are using the same methods.
Closed Cup vs. Open Cup.
 
Posts: 204 | Location: Invicta Oil Lab, Norway | Registered: Fri March 05 2004Reply With QuoteEdit or Delete MessageReport This Post
Platinum Member - 50 or more posts
Posted Hide Post
The open cup method is not recommended for those materials whose flash point is less than 75°c as per the ASTM methods.(ASTM D 92).Also this much difference is not possible.
 
Posts: 90 | Registered: Thu December 16 2004Reply With QuoteEdit or Delete MessageReport This Post