I would like to ask if anybody has an experience with laser particle counters used for analysis of fully synthetic diesel engine oils (i.e. 0W30)? We want to find and use a laser particle counter for such oil contamination monitoring. The target is an identification of acceptable contamination limits. Our concern is that such oils can be simply 'too black' for laser analysis and additives could distort the measurement...
Do you have any experience? Thank you in advance.
Original Post

Replies sorted oldest to newest

The Promlem with engine oil laser particle counting is twofold.
1/engine oils are dark, especially after use
2/ large particles will block the sensor and cause problems.
We have done this work in the past but found the results to be unsatisfactory(very high count), high dilution factorr required and sensor blocking
Other methods ie pore blocking only give an ISO code estimate.
Not tried patch test with auto counting which may be the answer.
Both Laser particle counters (>4 um) and older style light bulb particle counters all operate on the light blockage method. Neat in service crankcase oils are too dark to permit crisp definition of particle sizes of interest. There are three solutions. The first is to determine the % soot by some means. if the soot is <<0.1% then you can try dilution with a compatible solvent (ask your lubricant supplier.) Ihis may not work. If this does not work, one alternate is ASTM F312 Standard Test Methods for Microscopical Sizing and Counting Particles from Aerospace Fluids on Membrane Filters. This is very accurate but will require a qualified lab. The last, and unfortunately the preferred method is the pore blockabe method. This method will ignore the soot and count only the hard particles. Unfortunately there is no calibration standard for this method that I know of. Good luck
amigo Andrew los valores de niveles de limpieza ISO (1999 ) 4/6/14 um se distorcionan mucho y sobre todo el mas afectado es el conteo de particulas sliding, cuando la muestra de aceite de motor presenta fibras en exeso ,el laboratorio dode laboro ha realizado una investigacion para saber las larmas que nos permiten saber que numero de fibras no afecta el iso ni la repetibilidad del analisis,cuando repetir el analis pues hay casos cuando la cantidad de fibras es moderado en un segundo analisis no se detectan fibras.
Pero en el caso de altas concentraciones de fibras un analisis exacto y presiso es imposible.
Por favor comuniquese con migo para darle mayores detalles .nehemiasweb@hotmail.com
Post
×
×
×
×
Link copied to your clipboard.
×