Couple of points

Advantages of Spec M
-Spec M requires no sample prep (dilution).
-has no running costs if you don't need it.
-Volatises 0-8µm particle sizes (more range than ICP)

Advantages of ICP
-Has the element Sulphur - less important these days with low sulphur diesel.

Cons of ICP
-High running costs if no samples done - constantly needs to be on, and while on consumes gas.
-can only volatilise samples between 0-4µm sizes.

-------

If you do less than 3000 samples a month go for Spec M definitely. Cost of each sample = electrode sharpened + electrode disc

= Rand 3.00 = < $0.30c per sample (probably cheaper if your in the US).

Only conside the ICP if you do over 3000 samples a month - it has high running costs associated with the fact it constantly uses gas even when not analysing samples. However the cost per sample is minimal.

Andrew
If I was more interested in wear metals, definitly the Spectro. Spectro has these units set up for the right range of wear metal interest (we have two of them, one for RDE and one for RFS). Running costs are definitely acceptable. Spare parts, after-sales service and upgrades have been outstanding.

If I was more interested in additive levels, or had a significant additive level concern, probably an ICP, since it is linear through many orders of magnitude, and oil additives can range from very low to very high. I have only heard of the high running costs, and we have avoided ICP due to those stories (heard one from a Cat SOS lab themselves).

Also heard a rumour, and it may be only this, a rumour, that the reason Cat set ICP as their standard so long ago, even though the particle size issue was known, was that at the time, the Spectro had no automation, which it now has.

The Spectro can do additives, but you can run into ppm disagrements with oil company labs who are usually using ICP's. When we have a sample that is being analyzed for additive interest, we make it known to the customer that there is the possibility that our additive results may differ from the oil company results or other lab results that are using an ICP.
quote:
Originally posted by Marat Turumbetov:
Does anyone have experince with usage SPECTROIL M/C-W Oil Analysis Spectrometer in CAT SOS labs? What about it's real advantages and disadvantages in comparance with ICP?

Regards,
Marat
didicoi_2004@mail.ru

ICP is absolutely the best and most accurate and reliable instrument. It spits out at once concentrations of additives, contaminants, and wear. I guess all those folks that constantly talk about particle size limitations of ICP (0-4µ, or 0-10µ, or whatever) probably never heard of acid reflux digestion (not a rocket science at all). Once an aliquot of oil sample is digested, all metals are in solution, the solution is analyzed by ICP and concentration determined, and all that talk of size limitation becomes a bunch of bull (what actually is for everyone familiar with the capacity of ICP instrument). And here we are talking about capacity of accurately analyzing dozens and dozens of different metals, like no other method, period. Of course, just like with everything else, one needs to run a baseline (new oil) in the same manner to be able to take the full advantage of the complete and accurate picture of all metal in their oil.
Bob K.,

During my employment with a couple of prominent commercial laboratories, and later at the USACE North Pacific Division Materials Lab, we mostly used hot conc. nitric acid reflux digestion, which is tremendous oxidant, and was sufficient to address clients/Corps’ requests.

Anyhow, I am not trying to be argumentative here, but I am really interested to know, how, and to what degree of certainty, you guys measured 1-3 micron particles in establishing (ICP particles limitation”), when there was not a single reliable standard known for that range of particles, until the ISO 4406 in 1999 introduced first traceable and reliable/repeatable 11171 calibration standards? Also, did you, or do you know if anyone else confirmed those “limitations" by using the traceable ISO 11171 standard levels? There are supposed to be limitations for RDE too, but I don’t see any comment to that extent. All I see is ICP technique being constantly marginalized when talking about trace metals. Could it be that there are some other reasons for downplaying ICP and favoring RDE that has nothing to do with science but everything to do with profit making?
Thanks Bob. This was quite an interesting (and logical) endeavor. I admire people who have the knowledge and strong attitude to find answers, and are willing and able to elaborate their achievements (failures too, as they are part of the learning curve). I see your interest is, and was, in advancing practical application of science, and your outlook seems untainted by any outside influence. I wish we had crossed paths at some point in our professional careers, as I always had similar predisposition and curiosity (being devil’s advocate is just one little segment of it). Cheers!