Hi everybody, I´m Spectroil MC-W user, so I would like to share some information with us, I have some problems at the moment to burn used oil like Caprinus, Gadinia and Arginas, those lubricants mainly the Arginas have high Copper content, and increse the content of Sodium, for example, I Burn an Argina XL40, the result in Cu is out of limit of the spectroil in Cu, the result is CU = ~7520 ppm, and Na= 173 ppm, so I think the Cu make a interference in Na. So as a solution I disolve the sample with base oil 0 ppm, 1 to 5, and the new result is CU 2264 ppm and Na = 7 ppm, the multiplying this result per 5 gives Cu 12260, and Na 35, Could you help me pls?, PD: the equipment was calibrated.
Original Post

Replies sorted oldest to newest

JC,

I stopped played with these toys approx. 20 yrs ago, but as far as I remember, not all elements’ concentrations when rerun diluted will mathematically correctly agree with the straight run concentrations; especially not if strong interferences are present that cause false-positive or false-negative results.

It is possible that Cu may have caused false-positive reading on Na. However, if you additionally made 1/2 and 1/10 dilutions and run them, you would have more info to better assess the degree of the interferences and report correct results. Additional runs of spiked straight and diluted samples with Na and Cu would help you reveal if there was any interference with Na result or how much it affected the results.

Also, it would be good to know what concentrations of Cu and Na you used for calibration, and what the detection limits and max acceptable reporting limit for those two elements are, after which you need to make dilutions.

I don’t think it is a good idea to dilute your sample with any oil as long as it does not contain Cu. Dilutions should be made with solvents or oils standardized for this purpose.

I wish you noted other trace metals in this sample and how their straight run results agreed with results of the 1/5 dilution.
Last edited by johnm
Another aspect to consider - I assume your spectrometer is using a rotating disk electrode - the oil film produced on the rotating disk surface is strongly affected by the sample's composition,viscosity, other properties. So if the oil you are diluting with doesn't match the composition and viscosity of the used oil, then the determined concentration x the dilution factor won't equal the undiluted sample.
Do you know the linearity range for Cu on this spectrometer? If you are outside the linear range, further discussions are not really useful.